| | #61 (permalink) |
| All Star Join Date: Feb 2006 Location: Oceanside, CA
Posts: 73
| Paul, You are relying on a part number to bring legitimacy to your product. We rely on Clinical data and trials...like all legit medical manufacturer's do. I have never seen any medical papers or studies written about ANY of your products....Why not?? You produce a clinical study using YOUR piece of equipment then we can talk. Looking forward to meeting you at ITA this year!! Should be interesting. Good Day. |
| | |
| | #62 (permalink) | |
| All Star Join Date: Feb 2006 Location: Oceanside, CA Posts: 73 | Quote:
| |
| | |
| | #63 (permalink) |
| All Star Join Date: Feb 2006 Location: Oceanside, CA Posts: 73 | Solar Steve, Can you please figure out the reading from the LTS device showing below. Following the math you have here I came up with a reading of 1.2 J/cm2....is this right? Thank you for your help. Cheers |
| | |
| | #64 (permalink) |
| UV Geek Squad Join Date: Nov 2004 Location: Lake St Clair Posts: 3,274 | Paul.... you blew the meter off it's 199.9 mW/cm² scale. That's a "bit" of improvement over your first units ~2 years ago reading like about 10! Hold the meter at "face distance" away from the LED array (about 2" ?) and hopefully the reading will show on the LCD. Todd.... That "1" display on the meter means "over-range". The full scale 199.9 mW/cm² for the same 10 minutes would be 0.199 W/cm² x 600 sec = 119.9 J/cm² fyi. __________________ ![]() ......................Flashback 2001........................ "One of the 'ORIGINAL' TanToday Gang" |
| | |
| | #65 (permalink) |
| Rookie Join Date: Nov 2004 Posts: 20 | The fact is that there is 20 years of medical research science behind the Omnilux LED Light Therapy panels along with a 510K clinical trial and FDA OTC clearance based on those clinical trials. There are many factors that contribute to the effectiveness of LED Photo Therapy some of those being total energy output, distance from subject, nanometer levels, what different nanometer levels are transmitted, in what sequence the skin is subjected to the different nanometer levels, time between exposure to the different nanometer levels and more. For those of you that are looking for a more indepth understanding of the science behind LED Light Therapy I suggest you log on to Omnilux : Leaders in light therapy : Home and you can go to Tanning beds and commercial tanning beds by ETS to check out the facts on the Rejuvasun Spa bed with Omnilux Light Therapy. From an industry perspective this new technology opens many opportunities for revenue increases, new customers and a new service that even the Derms will have a hard time complaining about as long as the manufacturers are using proven technology that utilizes the same science/technology as the LED Phototherapy units that they currently use. At last count over 3000 doctors are providing LED Photo Therapy sessions with Omnilux devices. |
| | |
| | #66 (permalink) |
| Rookie Join Date: Nov 2004 Posts: 20 | Website Address for Omnilux Omnilux : Leaders in light therapy Website for Rejuvasun Spa Bed Tanning beds and commercial tanning beds by ETS |
| | |
| | #67 (permalink) | ||||||||||
| UV Geek Squad Join Date: Nov 2004 Location: Lake St Clair Posts: 3,274 | Per Bill's website link... the specs are: Specifications Omnilux reviveTM
Users of the Solarmeter 9.6 meter have in the past posted readings around 100 mW/cm² for Omnilux and near 80 mW/cm² for Lumiere fyi. __________________ ![]() ......................Flashback 2001........................ "One of the 'ORIGINAL' TanToday Gang" | ||||||||||
| | |
| | #68 (permalink) | |||||||||||
| Smartass Canuck Moderator Join Date: Feb 2005 Location: Canada Posts: 8,888 | Quote:
What would ITS readings show....you mentioned above 119.9 in a 10min session. Does that change for a 20min session time? I don't understand the differences between the readings. __________________ ![]() www.ChronosMarketing.com ____________________________________________ The problem nowadays is stupidity. Why don't we just take the safety labels off everything and see what happens? | |||||||||||
| | |
| | #69 (permalink) |
| UV Geek Squad Join Date: Nov 2004 Location: Lake St Clair Posts: 3,274 | It would be twice the 119.9 = 239.8 J/cm² for 20 minutes.... IF it read 199.9 mW/cm² at "face position"... distance from LED cluster surface to face surface (cheeks not nose). We don't know what it reads at that distance because he stuck the meter right up against LEDs it looks like in picture. Don't confuse meter readings with dose. Dose is intensity x time. __________________ ![]() ......................Flashback 2001........................ "One of the 'ORIGINAL' TanToday Gang" |
| | |
| | #70 (permalink) |
| Rookie Join Date: Jan 2008 Location: Blairsville, PA Posts: 15 | So the ITS unit should get the same results in less time? I'm thinking of adding a unit to my salon and it seems like everyone is bashing the ITS unit but yet it seems to be the one making the most power. |
| | |
| | #71 (permalink) |
| UV Geek Squad Join Date: Nov 2004 Location: Lake St Clair Posts: 3,274 | Seems-to-be and IS are two different things. Need to see a completed array reading at proper position. The readings should not be via sticking the meter sensor right on top of an LED. That's like taking a high pressure UV lamp reading "at the glass" instead of at body and/or face position. __________________ ![]() ......................Flashback 2001........................ "One of the 'ORIGINAL' TanToday Gang" |
| | |
| | #72 (permalink) |
| Rookie Join Date: Jan 2008 Location: Blairsville, PA Posts: 15 | My apologies. I guess what I'd like to see is some meter readings from the Lumiere machine. I could care less about part numbers I have no way of looking up, I want to see how much power both machines put out. All I've seen so far, is ITS posting pictures, even if the meter is too close, it still shows something right? |
| | |
| | #73 (permalink) |
| UV Geek Squad Join Date: Nov 2004 Location: Lake St Clair Posts: 3,274 | The meter readings and pics from Lumiere and Omnilux were discussed and posted about 2 years ago (Circa June 2006). I will email the people who did so and ask if they want to join this thread with data and images. You can try a "search" of the site in meantime if you wish. __________________ ![]() ......................Flashback 2001........................ "One of the 'ORIGINAL' TanToday Gang" |
| | |
| | #75 (permalink) |
| UV Geek Squad Join Date: Nov 2004 Location: Lake St Clair Posts: 3,274 | Well.... as I recall - the readings were taken 1-2 inches from LED array (maybe 1" from nose or 2" from cheek). The Omnilux read higher than Lumiere mostly because there were not any divider-bars between the clusters of LEDs (I think). As with tanning beds for UV readings... illumination geometry is important (canopy curvature, reflectance, etc), and of course amount of LEDs plus how closely packed they are. Also input wattage and yes individual LED output in the 633 ±xx nm bandwidth SPD narrow bell curve vs meter responsivitity: ![]() Got all that? Does Underhill lol? I have NO foggy idea what the TA or ProSun red LED units put out. __________________ ![]() ......................Flashback 2001........................ "One of the 'ORIGINAL' TanToday Gang" |
| | |
| | #76 (permalink) |
| Hall of Famer Join Date: Dec 2004 Location: Cloud 9 Posts: 1,353 | Steve, I would love to post some readings, but I do not have access to post pictures in the gallery. I have been following this thread for some time now, but have opted to not put in my 2 cents, but I can not hold back any longer. As far as readings go, I have metered several Lumiere units and one Omnilux. The Lumieres I metered vary from 78 mw/cm2 to 84mw/cm2, the Omnilux was @ 96mw/cm2 I metered an ITS unit in Vegas and it only hit 12mw/cm2. The guy running the booth told me (only after) the reading that it was not at full power. So I told him to turn it up; he said no, I can’t do that. Ummm BS! We can all do the math on that one. I don’t care how strong your LED’s are Paul, I care about actual dose to the face. The Lumiere unit has something like 2100 LED’s. How many are you working with? What is the reading off your unit at a practical 2-3 inches? BTW: I do hope you finally got some power in that thing. FYI: Needs a new look, to “Battlestar Galactica” for most salons/spas. IMHO __________________ "We are what we repeatedly do. Excellence, then, is not an act, but a habit." - Aristotle "The first step to becoming is to will it." - Mother Teresa |
| | |
| | #77 (permalink) |
| Hall of Famer Join Date: Dec 2004 Location: Cloud 9 Posts: 1,353 | ETS Tanning/633 combo… I do not believe in it at this point. 1.) Panel is too far from face to be at peak effectiveness. An LED’s power fades very fast. 2.) Cross contamination (do not know how else to put it) of other rays, specifically UVA I remember Tony showing me a note from PTL that stated unequivocally that Lumiere customer should in no way tan their face and that any use of a device emitting UV would be counter productive. So why the sudden change in opinion by PTL? Is this Dr. Whitehurst opinion too, or just the marketing teams opinion? Fact is that PTL does have the due credit in this field. They discovered it, they perfected it and now they get knocked off by others. ETS partnered up with the right company, but I would like to know about the 2 issues raised. IMHO __________________ "We are what we repeatedly do. Excellence, then, is not an act, but a habit." - Aristotle "The first step to becoming is to will it." - Mother Teresa |
| | |
| | #78 (permalink) |
| Hall of Famer Join Date: Dec 2004 Location: Cloud 9 Posts: 1,353 | Now for the kicker! What about 830nm? According to another experienced (past) Lumiere operator, he states that the magic is really in the 830nm when used with the 633nm as per the general protocol for facial rejuvenation. He found this out when he sold his Lumiere to purchase an Omnilux with the 830 head. Does 633 work with out the 830? Yes, I have plenty of B&A pictures to prove it to myself, staff and clients. However, the boost in facial regeneration is heavily amplified by the combo with 830nm. So who am I to speak on this? We operate a Lumiere unit, one of the first in the market and have done over 5,000 sessions w/ hundreds of clients. My opinions are my own and based in real hands on experience for 4+ years of offering this service. FYI: The B&A picture in the ad (page 14, IST; April 2008) for the rejuvasun332 bed is from a salon in Florida IMHO __________________ "We are what we repeatedly do. Excellence, then, is not an act, but a habit." - Aristotle "The first step to becoming is to will it." - Mother Teresa |
| | |
| | #79 (permalink) |
| Hall of Famer Join Date: Dec 2004 Location: Cloud 9 Posts: 1,353 | PTL = Tested, Proven, Clinical Proof, Leader ITS = New, w/Hand Treatment*, Competition for PTL, May be Stronger**
IMHO __________________ "We are what we repeatedly do. Excellence, then, is not an act, but a habit." - Aristotle "The first step to becoming is to will it." - Mother Teresa |
| | |
| | #80 (permalink) | |
| Rookie Join Date: May 2007 Location: Indianapolis Posts: 14 | Quote:
1. PTL measured their dose at the correct distance that a human face would be from the LED panel. This is the same dose that was utilized in clinical studies. 2. There is no cross contamination of UV rays as the HP facials cannot be activated at the same time as the LED panel. 3. The ETS Rejuvasun bed utilizes 633nm and 830nm LED's in alternating sessions as per the clinical study. | |
| | |
![]() |